I did a quick upgrade of my one today, and added the input protection it should have had ( on the PCB but omitted for cost reasons) by adding 4 275V varistors in the required locations. I used 2 in series for each poswition, so now the input will be clamped for any voltage over 830V DC peak, as I selected the varistors for equal breakover voltage. I found a pack of RS 204-1996 275VAC varistors i had ordered long ago, and checked all 5 to see breakover voltage. 2 were 402/3V, and 2 were 425/7V, with the one being 411V, so I used the high and low in series, so each chain has a similar breakdown voltage. Did not dismantle the meter, just soldered the pretinned and precut and preformed ends into the holes in the PCB, and then joined the midpoints of the 2 strings, so that they were floating in the air inside. Placed the varistors neatly in the space and closed it up again.Checked on resistance, and still reads open circuit, so no problems there. Was debating doing a high pot test but as they have been tested before use decided against it, they should work. I know the little green tester from Frankie can deliver 1400V as a test voltage, but I still like the meter. Ohers have tested them to destruction, and the PCB breaks down at around 1500V so I am happy enough.