Hi Martin,
Looking at the manual which I found at
http://www.aidetek.com/New_products_info/Datasheet/huayi/MS5308_english_manual.pdfthe ranges do not only tell us about resolution and accuracy. We may also notice, that many of the capacities you tested are far out of the described ranges.
4000uF, 100uF seems to fit only the lower scales.
Highest Capacity Ranges at different frequencies:
Frequency Capacity Res. Accuracy
100Hz/120 20.00mF 0.1mF 2.0%+5d # This is within spec
1kHz 2000.0uF 0.1uF 1.0%+5d # Already out with 4000uF but ok for 95uF
10kHz 2000.0nF 0.1nF 2.0%+5d # Far over the limit
100kHz 200.00nF 0.01nF 5.0%+5d # Way beyondLet's take a look at the expected values of the
reactance for the desired capacities and frequencies:
X(C,f) = 1/(2*pi*C*f)
C f X
95uF 100Hz 16.75 Ohm # OK
95uF 1kHz 1.68 Ohm # still good
95uF 10kHz 0.17 Ohm # a little low
95uF 100kHz 0.02 Ohm # getting tiny@min 26:00 The issue is that you are 100 times beyond the spec.
Testing 95uF at 100kHz does not makes any sense to me.
@min 28:36
Wow, now you go even further!
C f X
4.3mF 100Hz 0.37 Ohm
4.3mF 1kHz 37 mOhm
4.3mF 10kHz 3.7 mOhm
4.3mF 100kHz 0.37 mOhmEven the pure resistance test at 100kHz has only a resolution of 1mOhm:
100kHz 20.000Ohm 0.001 Ohm 1.0%+5d0.4 mOhm is way beyond what makes sense to measure.
100Hz is in spec, but every thing above is just good will.
There are not enough "pokie out tongues" to express the relation of imaginary 0.37 mOhm to the real world of electronics

@min 31:00
1uF is also not such a small capacitor. It is out of spec at the 100kHz frequency.
@min 32:00
10uF? Seriously? You should go down instead of up with the capacity if you would like to make sense at higher frequencies.
@min 33:45 "high losses in a 10uF at 10kHz" Who cares? Dissipation and quality factors matter for oscillators, which use much smaller capacities to achieve high frequencies.
@min 34:23 "heavens"
@min 34:40 "that shows 44, which is really high" no it is not, it is
nF not uF, but at 100kHz this is all just house numbers.
"2.48 which is really low", no 2.48
uF= 2480
nF. You are confused. I sometimes need mind overload protection, when viewing your videos

@min 35:00: it is again way beyond the spec, which ended at 2uF / 10kHz.
Measurement at 1kHz would be OK, but not at 100kHz.
@min 37:
Phase angle is an interesting option. You may try this with your high capacity high frequency mysteries. There, the phase angle should dramatically drop to show you, how far you get away from the ideal (-90°) capacitor.
4.74 Ohms ESR and
C = 2.48uF @ f= 100kHz
makes X(C,f) = 1/(2*pi*f*C) = 0.641 Ohm pretty low compared to the ESR!
atan(Xc/ESR)= 7,71° Getting far beyond the
cutoff frequency (phase angle 45°) even with the ESR makes no sense to me.
@min 39:30 ESR and capacity are probably not measured at the same frequency. A large capacity at high frequency shortens the AC current and drops the AC reactance (Xc) very low, which is fine to measure the ESR. But it is no good to measure capacity.
Regards,
Stephan
PS.: Before making the tutorial about capacitors, you may dig a little deeper into the theory of them. Take a look at
Jeri Ellsworth's Vimeo Video about capacitors. But don't kill any birds with them

By the way, your analogy to balloons and bladders is opening Pandora's box in term of your hydraulic example. Once you open the tubes with
monopoles where there should be only
dipoles, the electric sauce may flow out uncontrollable

The capacitor should be seen like all other elements as having at least two pipes connected to the circuit. You can regard the capacitor as a vessel with an elastic membrane in it.