The protection should not be subject only to fuse rating but also to PCB track layout and other factors. The conformity should not be only to fuse rating. The conformity should be checked by the multifunction tests of IEC61010-1 2nd Edition Clause 16.2 as followings.
---------------------------------------------------------------------------------------------------------------------------------------------
Multifunction meters and similar equipment shall not cause a HAZARD in any possible combination of RATED input voltages, and settings of function and range controls. Possible HAZARDS include electric shock, fire, arcing and explosion.
Conformity is checked by the following test.
The maximum RATED voltage specified for any function is applied to each pair of TERMINALS in turn, in every combination of function and range controls. The test source connected to the equipment measuring TERMINALS during this test is limited to 3,6 kVA for measurement category I or measurement category II. For measurement category III or measurement category IV, the test circuit has to be capable of delivering 30 kVA.